• ICP-OES

    From jmsm357@gmail.com@21:1/5 to Inor on Sun Oct 18 10:31:37 2015
    On Friday, December 12, 2003 at 10:51:21 AM UTC-7, Inor wrote:
    Hello,
    I've been running ICP-OES for three years now. The Instrument is a PerkinElmer Optima 3000. It's about ten years old. We bought it used in 1995 or 96. I've been struggling with what seems to be an ever-worsening sensitivity/ low-end stability problems. Most notably, when I run a low-level (10 ppb) check on the lower wavelength elements, i.e As, Se, and Tl, I have a hell of a time getting them to pass. We've replaced mirrors, lenses, PA tube and the sample intro system is cleaned/replaced every three months or so. My question is, do these things eventually just wear out (it gets alot of use, environmental monitoring: soils and waters). Are there
    any suggestions short of buying a new instrument (although I know some of
    the new ones are capable of sub-ppb levels on some elements). Any help is greatly appreciated.
    R. Lawrence
    Paradigm Analytical Labs

    High! I ran an Optima 3000 for a few years, but it was eventually replaced by a Varian. I liked the Optima, but the autosampler really sucks. Have you tried looking at the wavelength profile for the elements you're having problems with and examined the
    background base line? I think this technique is very useful when your dealing with determining concentrations close to the LDL.