• Measuring Extinction Ratio of a fiber laser

    From alex@21:1/5 to All on Sun Jan 13 12:40:47 2019
    I am getting poor polarization measurements from a fiber laser made from PM maintaining fiber. Firstly I would like to ask if my measurement method is ok.

    I collimate the beam with an aspherical lens of 8mm focal distance. Then there is a Glan Tayler Polarizer.
    https://www.thorlabs.de/newgrouppage9.cfm?objectgroup_id=815

    I tune the angle for maximum and minimum transmission, noting the power. I am only getting a 20:1 ratio between minimum and maximum transmission.

    I am concerned that my minimum measurement is my big source of error..Any change in this makes a large change in ratio. I am using a low and high power powermeter for each of the measurements for best accuracy.

    Is there another way of doing this measurement?
    Thanks
    Alex

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From JTS@21:1/5 to All on Sun Jan 13 23:06:39 2019
    Am 13.01.2019 um 21:40 schrieb alex:
    I am getting poor polarization measurements from a fiber laser made from PM maintaining fiber. Firstly I would like to ask if my measurement method is ok.

    I collimate the beam with an aspherical lens of 8mm focal distance. Then there is a Glan Tayler Polarizer.
    https://www.thorlabs.de/newgrouppage9.cfm?objectgroup_id=815

    I tune the angle for maximum and minimum transmission, noting the power. I am only getting a 20:1 ratio between minimum and maximum transmission.

    I am concerned that my minimum measurement is my big source of error..Any change in this makes a large change in ratio. I am using a low and high power powermeter for each of the measurements for best accuracy.

    Is there another way of doing this measurement?
    Thanks
    Alex



    I do not know what is the best way of doing this measurement, but I do
    know how to test your method: measure the polarization of light directly
    out of the laser (attenuated if necessary). If it works for that, then
    the method is ok.

    BTW a detector with a measurement range of 10^6 is standard (I do not
    know the reason why one can get such a wide range, I took the datum for
    one of the Thorlabs detectors), so it is likely that the method will be
    ok. Perhaps stray light???

    For alternative methods, hopefully someone else will give you some
    suggestions.

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From JTS@21:1/5 to All on Sun Jan 13 23:07:51 2019
    Am 13.01.2019 um 23:06 schrieb JTS:


    I do not know what is the best way of doing this measurement, but I do
    know how to test your method: measure the polarization of light directly
    out of the laser (attenuated if necessary).


    Here I mean: of *a* laser whose polarization you are sure about.

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From alex@21:1/5 to JTS on Sun Jan 13 14:22:09 2019
    On Sunday, January 13, 2019 at 11:07:52 PM UTC+1, JTS wrote:
    Am 13.01.2019 um 23:06 schrieb JTS:


    I do not know what is the best way of doing this measurement, but I do
    know how to test your method: measure the polarization of light directly out of the laser (attenuated if necessary).


    Here I mean: of *a* laser whose polarization you are sure about.

    Thanks for the idea... I will give that a go tmrw, but I think it might be harder tha it should be to find a laser whose polarisation we know... I'll let you know...
    Thanks
    Alex

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Phil Hobbs@21:1/5 to alex on Mon Jan 14 13:53:19 2019
    On 1/13/19 3:40 PM, alex wrote:
    I am getting poor polarization measurements from a fiber laser made from PM maintaining fiber. Firstly I would like to ask if my measurement method is ok.

    I collimate the beam with an aspherical lens of 8mm focal distance. Then there is a Glan Tayler Polarizer.
    https://www.thorlabs.de/newgrouppage9.cfm?objectgroup_id=815

    I tune the angle for maximum and minimum transmission, noting the power. I am only getting a 20:1 ratio between minimum and maximum transmission.

    PM fibre is only good for ~20 dB extinction ratio, so you're not doing
    that badly.

    It's worse if the injected light isn't aligned with one of the fibre axes.

    Also note that the high birefringence of PM fibre means that it has a
    high temperature coefficient of birefringence, so that the polarization
    is very unstable, though it will stay generally aligned along one axis
    if it's launched correctly.


    I am concerned that my minimum measurement is my big source of error..Any change in this makes a large change in ratio. I am using a low and high power powermeter for each of the measurements for best accuracy.

    Is there another way of doing this measurement?

    What power level are you measuring, what's the beam diameter, and what
    meters are you using?

    Cheers

    Phil Hobbs

    --
    Dr Philip C D Hobbs
    Principal Consultant
    ElectroOptical Innovations LLC
    Optics, Electro-optics, Photonics, Analog Electronics

    160 North State Road #203
    Briarcliff Manor NY 10510

    hobbs at electrooptical dot net
    http://electrooptical.net

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From JTS@21:1/5 to All on Tue Jan 15 00:35:02 2019
    Am 13.01.2019 um 23:22 schrieb alex:
    On Sunday, January 13, 2019 at 11:07:52 PM UTC+1, JTS wrote:
    Am 13.01.2019 um 23:06 schrieb JTS:


    I do not know what is the best way of doing this measurement, but I do
    know how to test your method: measure the polarization of light directly >>> out of the laser (attenuated if necessary).


    Here I mean: of *a* laser whose polarization you are sure about.

    Thanks for the idea... I will give that a go tmrw, but I think it might be harder tha it should be to find a laser whose polarisation we know... I'll let you know...
    Thanks
    Alex



    I think Phil Hobbs' suggestions are more precise than mine, but another approach is - polarize the light with a polarizer (a polarizer with
    1:100 ratio is good enough for this) and repeat your measurements. If
    you can measure 1:100, it means the 1:20 you measured is ok.

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Phil Hobbs@21:1/5 to JTS on Mon Jan 14 23:33:43 2019
    On 1/14/19 6:35 PM, JTS wrote:
    Am 13.01.2019 um 23:22 schrieb alex:
    On Sunday, January 13, 2019 at 11:07:52 PM UTC+1, JTS wrote:
    Am 13.01.2019 um 23:06 schrieb JTS:


    I do not know what is the best way of doing this measurement, but I do >>>> know how to test your method: measure the polarization of light
    directly
    out of the laser (attenuated if necessary).


    Here I mean: of *a* laser whose polarization you are sure about.

    Thanks for the idea... I will give that a go tmrw, but I think it
    might be harder tha it should be to find a laser whose polarisation we
    know... I'll let you know...
    Thanks
    Alex



    I think Phil Hobbs' suggestions are more precise than mine, but another approach is - polarize the light with a polarizer (a polarizer with
    1:100 ratio is good enough for this) and repeat your measurements. If
    you can measure 1:100, it means the 1:20 you measured is ok.

    Yup. A Glan-Taylor is good for 1:10,000 if you use it right. Almost as
    good as a Wollaston, at least in the transmitted beam. A "beam
    splitting Thompson" improves the refracted beam a fair amount, but
    they're not that common. Stick with the transmitted one.

    Cheers

    Phil Hobbs

    --
    Dr Philip C D Hobbs
    Principal Consultant
    ElectroOptical Innovations LLC / Hobbs ElectroOptics
    Optics, Electro-optics, Photonics, Analog Electronics
    Briarcliff Manor NY 10510

    http://electrooptical.net
    http://hobbs-eo.com

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From giovanni.notebooks@gmail.com@21:1/5 to Phil Hobbs on Tue Jan 15 15:21:49 2019
    On Tuesday, January 15, 2019 at 5:33:51 AM UTC+1, Phil Hobbs wrote:


    Yup. A Glan-Taylor is good for 1:10,000 if you use it right. Almost as good as a Wollaston, at least in the transmitted beam. A "beam
    splitting Thompson" improves the refracted beam a fair amount, but
    they're not that common. Stick with the transmitted one.



    I became curious and I read a bit on Wikipedia. The reflected (I think you mean reflected, right?) beam is only partially polarized in these kind of polarizers because they work by total internal reflection of one polarization (which is then completely
    absent in the transmitted beam) but the other polarization is also reflected a bit (so it is present in the reflected beam).

    I have also read about the difference between Glan-Thompson (is this the one you called "beam splitting Thompson"?) and Glan-Taylor, but to figure it out I would have to spend some time with calculations, so I stop here for the moment.

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Phil Hobbs@21:1/5 to giovanni.notebooks@gmail.com on Wed Jan 16 10:45:18 2019
    On 1/15/19 6:21 PM, giovanni.notebooks@gmail.com wrote:
    On Tuesday, January 15, 2019 at 5:33:51 AM UTC+1, Phil Hobbs wrote:


    Yup. A Glan-Taylor is good for 1:10,000 if you use it right. Almost as
    good as a Wollaston, at least in the transmitted beam. A "beam
    splitting Thompson" improves the refracted beam a fair amount, but
    they're not that common. Stick with the transmitted one.



    I became curious and I read a bit on Wikipedia. The reflected (I think you mean reflected, right?) beam is only partially polarized in these kind of polarizers because they work by total internal reflection of one polarization (which is then completely
    absent in the transmitted beam) but the other polarization is also reflected a bit (so it is present in the reflected beam).

    The transmitted beam is the good one. In a Wollaston, both beams are
    equally good, and for a bonus there are no etalon fringes. Glan-Taylors
    make fringes because the transmitted beam enters and exits perpendicular
    to the facets, so there are multiple coaxial reflections.


    I have also read about the difference between Glan-Thompson (is this the one you called "beam splitting Thompson"?) and Glan-Taylor, but to figure it out I would have to spend some time with calculations, so I stop here for the moment.

    Normal Glan-Thompsons have black wax to absorb the reflected beam. It
    makes many bounces between the hypotenuse and the wax, so it gets
    absorbed very efficiently.

    Cheers

    Phil Hobbs


    --
    Dr Philip C D Hobbs
    Principal Consultant
    ElectroOptical Innovations LLC / Hobbs ElectroOptics
    Optics, Electro-optics, Photonics, Analog Electronics
    Briarcliff Manor NY 10510

    http://electrooptical.net
    http://hobbs-eo.com

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From roptics@gmail.com@21:1/5 to alex on Fri Mar 8 07:55:00 2019
    On Sunday, January 13, 2019 at 3:40:48 PM UTC-5, alex wrote:
    I am getting poor polarization measurements from a fiber laser made from PM maintaining fiber. Firstly I would like to ask if my measurement method is ok.

    I collimate the beam with an aspherical lens of 8mm focal distance. Then there is a Glan Tayler Polarizer.
    https://www.thorlabs.de/newgrouppage9.cfm?objectgroup_id=815

    I tune the angle for maximum and minimum transmission, noting the power. I am only getting a 20:1 ratio between minimum and maximum transmission.

    I am concerned that my minimum measurement is my big source of error..Any change in this makes a large change in ratio. I am using a low and high power powermeter for each of the measurements for best accuracy.

    Is there another way of doing this measurement?
    Thanks
    Alex

    Hello,

    What minimum power values are you seeing for the "min" measurement? The smaller the numbers the more you have to worry about ambient light and scatter etc. Also what is the structure of your fiber laser? The portion that you're measuring is it LMA or
    standard fiber and what is the NA?

    Robert

    --- SoupGate-Win32 v1.05