showing 2 copies of the expected display, one above the other.
Whether this is a bug or a feature is questionable but reseting
the scope to a common state would be helpful. Anyone know if this
is possible and how it's done?
In article <uqqpn6$asp$1@reader1.panix.com>, Hul Tytus <ht@panix.com> wrote: >>I have a Rigol DS1102E, a 2 channel 100mc scope, that is currently
showing 2 copies of the expected display, one above the other.
Whether this is a bug or a feature is questionable but reseting
the scope to a common state would be helpful. Anyone know if this
is possible and how it's done?
You have probably put the scope into "Delayed Scan" mode, where the
lower half of the window shows a magnified view of a portion of the
full sweep. You might have gotten into this mode via the horizontal
menu (press horizontal MENU, then select Delayed and then On) or by
pressing the horizontal scale button. The latter is probably quite
easy to do by accident.
You can turn this mode back off via the menu, and (I assume) by
pressing the horizontal scale button a second time.
There's also a split-screen mode if the FFT feature is enabled, but
that doesn't sound like what you're seeing.
I don't see a complete "system reset" feature anywhere in the manual;
I might be missing it. If it's there it's probably buried in the
"Utility" menus somewhere.
In article <uqqpn6$asp$1@reader1.panix.com>, Hul Tytus <ht@panix.com> wrote: >I have a Rigol DS1102E, a 2 channel 100mc scope, that is currently
showing 2 copies of the expected display, one above the other.
Whether this is a bug or a feature is questionable but reseting
the scope to a common state would be helpful. Anyone know if this
is possible and how it's done?
You have probably put the scope into "Delayed Scan" mode, where the
lower half of the window shows a magnified view of a portion of the
full sweep. You might have gotten into this mode via the horizontal
menu (press horizontal MENU, then select Delayed and then On) or by
pressing the horizontal scale button. The latter is probably quite
easy to do by accident.
You can turn this mode back off via the menu, and (I assume) by
pressing the horizontal scale button a second time.
There's also a split-screen mode if the FFT feature is enabled, but
that doesn't sound like what you're seeing.
I don't see a complete "system reset" feature anywhere in the manual;
I might be missing it. If it's there it's probably buried in the
"Utility" menus somewhere.
On Sat, 17 Feb 2024 11:10:52 -0800, dplatt@coop.radagast.org (Dave
Platt) wrote:
In article <uqqpn6$asp$1@reader1.panix.com>, Hul Tytus <ht@panix.com> wrote:
I have a Rigol DS1102E, a 2 channel 100mc scope, that is currently >>showing 2 copies of the expected display, one above the other.
Whether this is a bug or a feature is questionable but reseting
the scope to a common state would be helpful. Anyone know if this
is possible and how it's done?
You have probably put the scope into "Delayed Scan" mode, where the
lower half of the window shows a magnified view of a portion of the
full sweep. You might have gotten into this mode via the horizontal
menu (press horizontal MENU, then select Delayed and then On) or by >pressing the horizontal scale button. The latter is probably quite
easy to do by accident.
You can turn this mode back off via the menu, and (I assume) by
pressing the horizontal scale button a second time.
There's also a split-screen mode if the FFT feature is enabled, but
that doesn't sound like what you're seeing.
I don't see a complete "system reset" feature anywhere in the manual;
I might be missing it. If it's there it's probably buried in the
"Utility" menus somewhere.
One very handy button on my Rigol is DEFAULT, which turns off all
sorts of stuff.
Thanks John, that sounds like what's needed. Where's it
located?
Hul
John Larkin <jl@997pothill.com> wrote:
On Sat, 17 Feb 2024 11:10:52 -0800, dplatt@coop.radagast.org (Dave
Platt) wrote:
In article <uqqpn6$asp$1@reader1.panix.com>, Hul Tytus <ht@panix.com> wrote:
I have a Rigol DS1102E, a 2 channel 100mc scope, that is currently
showing 2 copies of the expected display, one above the other.
Whether this is a bug or a feature is questionable but reseting
the scope to a common state would be helpful. Anyone know if this
is possible and how it's done?
You have probably put the scope into "Delayed Scan" mode, where the
lower half of the window shows a magnified view of a portion of the
full sweep. You might have gotten into this mode via the horizontal
menu (press horizontal MENU, then select Delayed and then On) or by
pressing the horizontal scale button. The latter is probably quite
easy to do by accident.
You can turn this mode back off via the menu, and (I assume) by
pressing the horizontal scale button a second time.
There's also a split-screen mode if the FFT feature is enabled, but
that doesn't sound like what you're seeing.
I don't see a complete "system reset" feature anywhere in the manual;
I might be missing it. If it's there it's probably buried in the
"Utility" menus somewhere.
One very handy button on my Rigol is DEFAULT, which turns off all
sorts of stuff.