• Using Geiger tube in proportional mode

    From Piotr Wyderski@21:1/5 to All on Wed Jan 19 15:56:46 2022
    Hi,

    the literature says that avalanche gas detectors are capable of being
    run in proportional mode where the voltage is below the Geiger plateau.
    Would it make sense to operate a sufficiently undervolted Geiger tube
    as a proportional detector for the purpose of simple gamma/beta
    spectroscopy or is there some inherent property preventing it? E.g. the
    gas composition?

    If it can, then would a 10-way monster like the SBT10A be partly run in proportional mode and partly in the usual avalanche?

    Best regards, Piotr

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Phil Hobbs@21:1/5 to Piotr Wyderski on Wed Jan 19 10:52:22 2022
    Piotr Wyderski wrote:
    Hi,

    the literature says that avalanche gas detectors are capable of being
    run in proportional mode where the voltage is below the Geiger
    plateau. Would it make sense to operate a sufficiently undervolted
    Geiger tube as a proportional detector for the purpose of simple
    gamma/beta spectroscopy or is there some inherent property preventing
    it? E.g. the gas composition?

    A proportional counter is a fairly heartbreaking thing to use for spectroscopy--even a sodium iodide scintillator is much much better.

    OTOH for a hobby project, it would probably work OK.

    If it can, then would a 10-way monster like the SBT10A be partly run
    in proportional mode and partly in the usual avalanche?

    ISTM that you'd need to prevent ions from migrating around inside
    somehow, unless you're okay with a fairly long dead time after each
    Geiger pulse.

    Best regards, Piotr

    Cheers

    Phil Hobbs

    --
    Dr Philip C D Hobbs
    Principal Consultant
    ElectroOptical Innovations LLC / Hobbs ElectroOptics
    Optics, Electro-optics, Photonics, Analog Electronics
    Briarcliff Manor NY 10510

    http://electrooptical.net
    http://hobbs-eo.com

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Piotr Wyderski@21:1/5 to Phil Hobbs on Wed Jan 19 17:15:43 2022
    Phil Hobbs wrote:

    A proportional counter is a fairly heartbreaking thing to use for spectroscopy--even a sodium iodide scintillator is much much better.

    OTOH for a hobby project, it would probably work OK.

    Yes, this is for an entirely hobby project. No great spectral resolution
    is required; in fact, anything resembling the right spectrum would be
    very rewarding. It is an off-label application driven by curiosity.

    ISTM that you'd need to prevent ions from migrating around inside
    somehow, unless you're okay with a fairly long dead time after each
    Geiger pulse.

    By necessity I am going to use only the readily available weak sources
    with pretty well-defined chemical composition: uranium and thorium,
    perhaps americium too. So the dead time should not be a problem.

    Best regards, Piotr

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Dimiter_Popoff@21:1/5 to Phil Hobbs on Wed Jan 19 19:43:27 2022
    On 1/19/2022 17:52, Phil Hobbs wrote:
    Piotr Wyderski wrote:
    Hi,

    the literature says that avalanche gas detectors are capable of being
     run in proportional mode where the voltage is below the Geiger
    plateau. Would it make sense to operate a sufficiently undervolted
    Geiger tube as a proportional detector for the purpose of simple
    gamma/beta spectroscopy or is there some inherent property preventing
    it? E.g. the gas composition?

    A proportional counter is a fairly heartbreaking thing to use for spectroscopy--even a sodium iodide scintillator is much much better.

    OTOH for a hobby project, it would probably work OK.

    If it can, then would a 10-way monster like the SBT10A be partly run
    in proportional mode and partly in the usual avalanche?

    ISTM that you'd need to prevent ions from migrating around inside
    somehow, unless you're okay with a fairly long dead time after each
    Geiger pulse.

    Best regards, Piotr

    Cheers

    Phil Hobbs


    Hi Phil, just related - have you experimented with NaI and some of the photodiodes etc. you seem to do a lot of work with? My current
    understanding is that PMT-s are still better by quite some margin
    but then I don't give NaI much thought (although our netMCA is
    HPGe grade we do have customers who use it with NaI though).

    ======================================================
    Dimiter Popoff, TGI http://www.tgi-sci.com ====================================================== http://www.flickr.com/photos/didi_tgi/

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Phil Hobbs@21:1/5 to Piotr Wyderski on Wed Jan 19 14:10:08 2022
    Piotr Wyderski wrote:
    Phil Hobbs wrote:

    A proportional counter is a fairly heartbreaking thing to use for
    spectroscopy--even a sodium iodide scintillator is much much better.

    OTOH for a hobby project, it would probably work OK.

    Yes, this is for an entirely hobby project. No great spectral resolution
    is required; in fact, anything resembling the right spectrum would be
    very rewarding. It is an off-label application driven by curiosity.

    ISTM that you'd need to prevent ions from migrating around inside
    somehow, unless you're okay with a fairly long dead time after each
    Geiger pulse.

    By necessity I am going to use only the readily available weak sources
    with pretty well-defined chemical composition: uranium and thorium,
    perhaps americium too. So the dead time should not be a problem.

        Best regards, Piotr

    "Lite salt" sold in the supermarket is half KCl, which is a decent test
    source too.

    Cheers

    Phil Hobbs

    --
    Dr Philip C D Hobbs
    Principal Consultant
    ElectroOptical Innovations LLC / Hobbs ElectroOptics
    Optics, Electro-optics, Photonics, Analog Electronics
    Briarcliff Manor NY 10510

    http://electrooptical.net
    http://hobbs-eo.com

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From John Larkin@21:1/5 to bombald@protonmail.com on Wed Jan 19 10:44:00 2022
    On Wed, 19 Jan 2022 15:56:46 +0100, Piotr Wyderski
    <bombald@protonmail.com> wrote:

    Hi,

    the literature says that avalanche gas detectors are capable of being
    run in proportional mode where the voltage is below the Geiger plateau.
    Would it make sense to operate a sufficiently undervolted Geiger tube
    as a proportional detector for the purpose of simple gamma/beta
    spectroscopy or is there some inherent property preventing it? E.g. the
    gas composition?

    If it can, then would a 10-way monster like the SBT10A be partly run in >proportional mode and partly in the usual avalanche?

    Best regards, Piotr

    Geiger tubes have a really tiny wire, to make very high e-fields close
    to the wire for lots of multiplication. I recall that ion chambers
    have more, bigger wires, or flat electrodes. That difference probably
    affects pulse-height spectroscopy.

    SBT10A is interesting, a multi-wire geiger tube. It would probably
    work OK in proportional mode, with maybe a little multiplier gain.

    Are the wires brought out to the connector individually? Looks like
    it.



    --

    If a man will begin with certainties, he shall end with doubts,
    but if he will be content to begin with doubts he shall end in certainties. Francis Bacon

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Dimiter_Popoff@21:1/5 to Phil Hobbs on Wed Jan 19 21:30:26 2022
    On 1/19/2022 21:20, Phil Hobbs wrote:
    Dimiter_Popoff wrote:
    On 1/19/2022 17:52, Phil Hobbs wrote:
    Piotr Wyderski wrote:
    Hi,

    the literature says that avalanche gas detectors are capable of being
     run in proportional mode where the voltage is below the Geiger
    plateau. Would it make sense to operate a sufficiently undervolted
    Geiger tube as a proportional detector for the purpose of simple
    gamma/beta spectroscopy or is there some inherent property preventing
    it? E.g. the gas composition?

    A proportional counter is a fairly heartbreaking thing to use for
    spectroscopy--even a sodium iodide scintillator is much much better.

    OTOH for a hobby project, it would probably work OK.

    If it can, then would a 10-way monster like the SBT10A be partly run
    in proportional mode and partly in the usual avalanche?

    ISTM that you'd need to prevent ions from migrating around inside
    somehow, unless you're okay with a fairly long dead time after each
    Geiger pulse.


    Hi Phil, just related - have you experimented with NaI and some of the
    photodiodes etc. you seem to do a lot of work with? My current
    understanding is that PMT-s are still better by quite some margin
    but then I don't give NaI much thought (although our netMCA is
    HPGe grade we do have customers who use it with NaI though).

    Nothing can touch a PMT for dark counts per unit area--they're five or
    six orders of magnitude better than an APD or SiPM / MPPC.

    A typical 100-um diameter APD has about the same dark count rate as a _four_inch_ PMT.

    Plus you gain another factor of at least 2 by using coupling gel between
    the crystal and the PMT face.

    You'd think it was right around 2 because you win etendue by the square
    of the refractive index, but it's actually better than that--the extra
    light at high angles gets totally internally reflected at the photocathod/vacuum surface, so it gets at least two passes through the
    PC, which raises the quantum efficiency considerably.

    There are tricks to do even better, e.g. by using prism coupling to make light rattle round inside the faceplate till it gets absorbed.  With a negative electron affinity (NEA) photocathode, that'll get you almost
    100% QE over a wide bandwidth.

    Cheers

    Phil Hobbs



    Thanks Phil.

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Phil Hobbs@21:1/5 to All on Wed Jan 19 14:20:19 2022
    Dimiter_Popoff wrote:
    On 1/19/2022 17:52, Phil Hobbs wrote:
    Piotr Wyderski wrote:
    Hi,

    the literature says that avalanche gas detectors are capable of being
     run in proportional mode where the voltage is below the Geiger
    plateau. Would it make sense to operate a sufficiently undervolted
    Geiger tube as a proportional detector for the purpose of simple
    gamma/beta spectroscopy or is there some inherent property preventing
    it? E.g. the gas composition?

    A proportional counter is a fairly heartbreaking thing to use for
    spectroscopy--even a sodium iodide scintillator is much much better.

    OTOH for a hobby project, it would probably work OK.

    If it can, then would a 10-way monster like the SBT10A be partly run
    in proportional mode and partly in the usual avalanche?

    ISTM that you'd need to prevent ions from migrating around inside
    somehow, unless you're okay with a fairly long dead time after each
    Geiger pulse.


    Hi Phil, just related - have you experimented with NaI and some of the photodiodes etc. you seem to do a lot of work with? My current
    understanding is that PMT-s are still better by quite some margin
    but then I don't give NaI much thought (although our netMCA is
    HPGe grade we do have customers who use it with NaI though).

    Nothing can touch a PMT for dark counts per unit area--they're five or
    six orders of magnitude better than an APD or SiPM / MPPC.

    A typical 100-um diameter APD has about the same dark count rate as a _four_inch_ PMT.

    Plus you gain another factor of at least 2 by using coupling gel between
    the crystal and the PMT face.

    You'd think it was right around 2 because you win etendue by the square
    of the refractive index, but it's actually better than that--the extra
    light at high angles gets totally internally reflected at the photocathod/vacuum surface, so it gets at least two passes through the
    PC, which raises the quantum efficiency considerably.

    There are tricks to do even better, e.g. by using prism coupling to make
    light rattle round inside the faceplate till it gets absorbed. With a
    negative electron affinity (NEA) photocathode, that'll get you almost
    100% QE over a wide bandwidth.

    Cheers

    Phil Hobbs


    --
    Dr Philip C D Hobbs
    Principal Consultant
    ElectroOptical Innovations LLC / Hobbs ElectroOptics
    Optics, Electro-optics, Photonics, Analog Electronics
    Briarcliff Manor NY 10510

    http://electrooptical.net
    http://hobbs-eo.com

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From John Larkin@21:1/5 to bombald@protonmail.com on Wed Jan 19 11:32:36 2022
    On Wed, 19 Jan 2022 17:15:43 +0100, Piotr Wyderski
    <bombald@protonmail.com> wrote:

    Phil Hobbs wrote:

    A proportional counter is a fairly heartbreaking thing to use for
    spectroscopy--even a sodium iodide scintillator is much much better.

    OTOH for a hobby project, it would probably work OK.

    Yes, this is for an entirely hobby project. No great spectral resolution
    is required; in fact, anything resembling the right spectrum would be
    very rewarding. It is an off-label application driven by curiosity.

    ISTM that you'd need to prevent ions from migrating around inside
    somehow, unless you're okay with a fairly long dead time after each
    Geiger pulse.

    By necessity I am going to use only the readily available weak sources
    with pretty well-defined chemical composition: uranium and thorium,
    perhaps americium too. So the dead time should not be a problem.

    Best regards, Piotr

    There are liquid scintillators too.

    --

    If a man will begin with certainties, he shall end with doubts,
    but if he will be content to begin with doubts he shall end in certainties. Francis Bacon

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Piotr Wyderski@21:1/5 to John Larkin on Wed Jan 19 21:12:07 2022
    John Larkin wrote:

    SBT10A is interesting, a multi-wire geiger tube. It would probably
    work OK in proportional mode, with maybe a little multiplier gain.

    Are the wires brought out to the connector individually? Looks like
    it.

    Yes, they are. It opens several interesting possibilities, e.g.
    shielding half of it to get beta and gamma signals separately or running
    some part of it in avalanche mode and the rest in proportional mode.
    Spectral sensitivity is also much better than typical metal tubes --
    some specs say it should react to gamma as low as 15keV.

    The digital part would be boring, just a bunch of counters. But it is
    the analog front-end that could make it possible to go way beyond the
    specs. Looks like a lot of fun in my spare time.

    Best regards, Piotr

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From whit3rd@21:1/5 to Piotr Wyderski on Wed Jan 19 13:37:34 2022
    On Wednesday, January 19, 2022 at 6:57:05 AM UTC-8, Piotr Wyderski wrote:
    Hi,

    the literature says that avalanche gas detectors are capable of being
    run in proportional mode where the voltage is below the Geiger plateau.
    Would it make sense to operate a sufficiently undervolted Geiger tube
    as a proportional detector for the purpose of simple gamma/beta
    spectroscopy or is there some inherent property preventing it? E.g. the
    gas composition?

    There's gas recipe differences, but that's because you want fat (time-duration) pulses for proportional but not Geiger. The big irritation, is that the proportional region is relatively small, it takes regulated V source, and
    only part of the (usually cylinder-around-wire) field region has best sensitivity.

    But, the construction, if not the gas mix and bias, are about the same.
    There's a lot of literature available online, like <https://archive.org/details/arxiv-1509.02379>
    and spectroscopy is possible... as long as your gas mix is dense enough to actually
    stop some of the rays in question. Some variant of multichannel analyzer is needed, of
    course, and a test source for calibration.

    Not sure about beta spectroscopy, unless you put the beta emitter inside the gas-containing tube.

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From John Larkin@21:1/5 to bombald@protonmail.com on Wed Jan 19 16:35:21 2022
    On Wed, 19 Jan 2022 21:12:07 +0100, Piotr Wyderski
    <bombald@protonmail.com> wrote:

    John Larkin wrote:

    SBT10A is interesting, a multi-wire geiger tube. It would probably
    work OK in proportional mode, with maybe a little multiplier gain.

    Are the wires brought out to the connector individually? Looks like
    it.

    Yes, they are. It opens several interesting possibilities, e.g.
    shielding half of it to get beta and gamma signals separately or running
    some part of it in avalanche mode and the rest in proportional mode.
    Spectral sensitivity is also much better than typical metal tubes --
    some specs say it should react to gamma as low as 15keV.

    The digital part would be boring, just a bunch of counters. But it is
    the analog front-end that could make it possible to go way beyond the
    specs. Looks like a lot of fun in my spare time.

    Best regards, Piotr

    I designed some wire chamber electronics for CERN. Each chamber was a
    big thing made of parallel sheets of aluminized mylar or something,
    with a zillion collector wires inside. I think there was some ion multiplication but not geiger mode. The wires were close enough that
    we could measure the pulses and interpolate the track locations
    between wires

    It was a massive flood of data from multiple planes, the object being
    to tease out particle tracks. We did "progressive enrichment" of the
    data: ecl gates, then FPGAs, then software. The final output was a
    single bit, PUBLISH.

    --

    If a man will begin with certainties, he shall end with doubts,
    but if he will be content to begin with doubts he shall end in certainties. Francis Bacon