• Ripple

    From Cursitor Doom@21:1/5 to All on Sun Feb 25 18:01:19 2024
    So - in the case of a linear power supply - the rectified output of
    the transformer typically goes straight to a big old storage cap for
    smoothing purposes. My question is: how much ripple should I expect to
    see across that cap if all's working well? I have to say I'm seeing a
    *lot* of ripple on this cap in a Marconi RF signal generator and it's
    finding its way to the RF output, which of course should ideally have
    zero ripple on it. The aforementioned cap tests fine WRT ESR and
    capacitance in-circuit, I should add.

    Thanks,
    CD.

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Cursitor Doom@21:1/5 to none@no.no on Sun Feb 25 18:54:48 2024
    On Sun, 25 Feb 2024 19:40:48 +0100, HW <none@no.no> wrote:

    On Sun, 25 Feb 2024 18:01:19 +0000, Cursitor Doom <cd@notformail.com>
    wrote:

    So - in the case of a linear power supply - the rectified output of
    the transformer typically goes straight to a big old storage cap for >>smoothing purposes. My question is: how much ripple should I expect to
    see across that cap if all's working well? I have to say I'm seeing a
    *lot* of ripple on this cap

    Check if a diode in the bridge rectifier has failed.

    Good idea! I must admit I'd overlooked that possibility. I should have
    checked the frequeny of the ripple whilst I was at it.

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From HW@21:1/5 to All on Sun Feb 25 19:40:48 2024
    On Sun, 25 Feb 2024 18:01:19 +0000, Cursitor Doom <cd@notformail.com>
    wrote:

    So - in the case of a linear power supply - the rectified output of
    the transformer typically goes straight to a big old storage cap for >smoothing purposes. My question is: how much ripple should I expect to
    see across that cap if all's working well? I have to say I'm seeing a
    *lot* of ripple on this cap

    Check if a diode in the bridge rectifier has failed.

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Liz Tuddenham@21:1/5 to Cursitor Doom on Sun Feb 25 19:16:24 2024
    Cursitor Doom <cd@notformail.com> wrote:

    On Sun, 25 Feb 2024 19:40:48 +0100, HW <none@no.no> wrote:

    On Sun, 25 Feb 2024 18:01:19 +0000, Cursitor Doom <cd@notformail.com> >wrote:

    So - in the case of a linear power supply - the rectified output of
    the transformer typically goes straight to a big old storage cap for >>smoothing purposes. My question is: how much ripple should I expect to >>see across that cap if all's working well? I have to say I'm seeing a >>*lot* of ripple on this cap

    Check if a diode in the bridge rectifier has failed.

    Good idea! I must admit I'd overlooked that possibility. I should have checked the frequeny of the ripple whilst I was at it.

    The shape of the ripple will tell you if that has happened or if one of
    the diodes haas gone open circuit. Another possibility is a fault in
    some other part of the circuit that is drawing too much current - you
    might smell something getting hot.


    --
    ~ Liz Tuddenham ~
    (Remove the ".invalid"s and add ".co.uk" to reply)
    www.poppyrecords.co.uk

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Roger Hayter@21:1/5 to Cursitor Doom on Sun Feb 25 19:52:55 2024
    On 25 Feb 2024 at 18:01:19 GMT, "Cursitor Doom" <cd@notformail.com> wrote:

    So - in the case of a linear power supply - the rectified output of
    the transformer typically goes straight to a big old storage cap for smoothing purposes. My question is: how much ripple should I expect to
    see across that cap if all's working well? I have to say I'm seeing a
    *lot* of ripple on this cap in a Marconi RF signal generator and it's
    finding its way to the RF output, which of course should ideally have
    zero ripple on it. The aforementioned cap tests fine WRT ESR and
    capacitance in-circuit, I should add.

    Thanks,
    CD.

    Usually there will either be a passive filter after the initial large capacitor, such as an inductor in series and a further large capacitor to earth, or an active voltage regulator circuit involving an amplifier and
    series semiconductor, or both. A simple diode plus capacitor is always going
    to have a huge amount of ripple, not hard to calculate if you know the current load. But no electronic instrument is going to be powered just by a rectifier capacitor combination. You need to look further into the circuit.


    --
    Roger Hayter

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Ken@21:1/5 to Cursitor Doom on Sun Feb 25 14:22:43 2024
    Cursitor Doom wrote:
    So - in the case of a linear power supply - the rectified output of
    the transformer typically goes straight to a big old storage cap for smoothing purposes. My question is: how much ripple should I expect to
    see across that cap if all's working well? I have to say I'm seeing a
    *lot* of ripple on this cap in a Marconi RF signal generator and it's
    finding its way to the RF output, which of course should ideally have
    zero ripple on it. The aforementioned cap tests fine WRT ESR and
    capacitance in-circuit, I should add.

    Thanks,
    CD.

    In addition to the other suggestions, you might check to see if a
    snubber capacitor exists across the diodes.

    https://www.sciencedirect.com/topics/engineering/snubber-circuit

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Cursitor Doom@21:1/5 to Liz Tuddenham on Sun Feb 25 23:59:36 2024
    On Sun, 25 Feb 2024 19:16:24 +0000, liz@poppyrecords.invalid.invalid
    (Liz Tuddenham) wrote:

    Cursitor Doom <cd@notformail.com> wrote:

    On Sun, 25 Feb 2024 19:40:48 +0100, HW <none@no.no> wrote:

    On Sun, 25 Feb 2024 18:01:19 +0000, Cursitor Doom <cd@notformail.com>
    wrote:

    So - in the case of a linear power supply - the rectified output of
    the transformer typically goes straight to a big old storage cap for
    smoothing purposes. My question is: how much ripple should I expect to
    see across that cap if all's working well? I have to say I'm seeing a
    *lot* of ripple on this cap

    Check if a diode in the bridge rectifier has failed.

    Good idea! I must admit I'd overlooked that possibility. I should have
    checked the frequeny of the ripple whilst I was at it.

    The shape of the ripple will tell you if that has happened or if one of
    the diodes haas gone open circuit. Another possibility is a fault in
    some other part of the circuit that is drawing too much current - you
    might smell something getting hot.

    There's nothing getting hot, but the ripple had a definite saw-tooth
    shape to it, which could be indicative of abnormal operation.
    Shame there are no schematics available anywhere on the net; always
    makes troubleshooting much harder.

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Liz Tuddenham@21:1/5 to Cursitor Doom on Mon Feb 26 09:59:12 2024
    Cursitor Doom <cd@notformail.com> wrote:

    On Sun, 25 Feb 2024 19:16:24 +0000, liz@poppyrecords.invalid.invalid
    (Liz Tuddenham) wrote:

    Cursitor Doom <cd@notformail.com> wrote:

    On Sun, 25 Feb 2024 19:40:48 +0100, HW <none@no.no> wrote:

    On Sun, 25 Feb 2024 18:01:19 +0000, Cursitor Doom <cd@notformail.com>
    wrote:

    So - in the case of a linear power supply - the rectified output of
    the transformer typically goes straight to a big old storage cap for
    smoothing purposes. My question is: how much ripple should I expect to >> >>see across that cap if all's working well? I have to say I'm seeing a
    *lot* of ripple on this cap

    Check if a diode in the bridge rectifier has failed.

    Good idea! I must admit I'd overlooked that possibility. I should have
    checked the frequeny of the ripple whilst I was at it.

    The shape of the ripple will tell you if that has happened or if one of
    the diodes haas gone open circuit. Another possibility is a fault in
    some other part of the circuit that is drawing too much current - you
    might smell something getting hot.

    There's nothing getting hot, but the ripple had a definite saw-tooth
    shape to it, which could be indicative of abnormal operation.
    Shame there are no schematics available anywhere on the net; always
    makes troubleshooting much harder.

    Is the rectifier circuit half-wave or full-wave? If the latter, is the
    ripple also full wave?

    Some designers allowed a larger amount of ripple on the first capacitor
    to reduce the heating of the mains transformer (the copper losses are proportional to I squared, so by allowing a longer charging time per
    cycle at a lower current, they achieved lower overall resistance
    losses). As long as the lowest part of the ripple was sufficiently far
    above the stabilised output, a series stabiliser could cope with it.

    Is the 'stabilised' line actually stable or is there ripple on it in the
    form of a series of little notches?


    --
    ~ Liz Tuddenham ~
    (Remove the ".invalid"s and add ".co.uk" to reply)
    www.poppyrecords.co.uk

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Roger Hayter@21:1/5 to Cursitor Doom on Mon Feb 26 10:51:03 2024
    On 26 Feb 2024 at 10:16:41 GMT, "Cursitor Doom" <cd@notformail.com> wrote:

    On Mon, 26 Feb 2024 09:59:12 +0000, liz@poppyrecords.invalid.invalid
    (Liz Tuddenham) wrote:

    Cursitor Doom <cd@notformail.com> wrote:

    On Sun, 25 Feb 2024 19:16:24 +0000, liz@poppyrecords.invalid.invalid
    (Liz Tuddenham) wrote:

    Cursitor Doom <cd@notformail.com> wrote:

    On Sun, 25 Feb 2024 19:40:48 +0100, HW <none@no.no> wrote:

    On Sun, 25 Feb 2024 18:01:19 +0000, Cursitor Doom <cd@notformail.com> >>>>>> wrote:

    So - in the case of a linear power supply - the rectified output of >>>>>>> the transformer typically goes straight to a big old storage cap for >>>>>>> smoothing purposes. My question is: how much ripple should I expect to >>>>>>> see across that cap if all's working well? I have to say I'm seeing a >>>>>>> *lot* of ripple on this cap

    Check if a diode in the bridge rectifier has failed.

    Good idea! I must admit I'd overlooked that possibility. I should have >>>>> checked the frequeny of the ripple whilst I was at it.

    The shape of the ripple will tell you if that has happened or if one of >>>> the diodes haas gone open circuit. Another possibility is a fault in
    some other part of the circuit that is drawing too much current - you
    might smell something getting hot.

    There's nothing getting hot, but the ripple had a definite saw-tooth
    shape to it, which could be indicative of abnormal operation.
    Shame there are no schematics available anywhere on the net; always
    makes troubleshooting much harder.

    Is the rectifier circuit half-wave or full-wave? If the latter, is the
    ripple also full wave?

    Marconi have used these things (not these exact components of course,
    but the same footprint/pinout:

    http://tinyurl.com/5ab3fd7h

    So full wave. Is the ripple full wave? Well, I did mean to check the frequency of it, but other stuff got in the way. I'm going to have to re-investigate a bit more thoroughly later in the week.

    Some designers allowed a larger amount of ripple on the first capacitor
    to reduce the heating of the mains transformer (the copper losses are
    proportional to I squared, so by allowing a longer charging time per
    cycle at a lower current, they achieved lower overall resistance
    losses). As long as the lowest part of the ripple was sufficiently far
    above the stabilised output, a series stabiliser could cope with it.

    Is the 'stabilised' line actually stable or is there ripple on it in the
    form of a series of little notches?

    Not sure what you mean by stabilised line. There are two secondary
    windings on the mains transformer which go into separate bridge
    rectifiiers of the type linked to above and then on downstream to the
    first tank caps. One line has a ton of ripple on it, the other very
    little indeed. Again, I'll need to go back and pay more attention to
    what's going on here. I've missed out too many important details.

    Older practice, when active devices were expensive and bulky was to have an active stabiliser circuit (which also got rid of the ripple) directly after
    the power supply. Nowadays they tend to have voltage regulators for each
    module or circuit, local to that module. If you have the latter arrangement I would be looking for ripple on the output of each local regulator rather than the main power supply. Or the high ripple could be due to excessive current being drawn from one power supply. Anything getting excessively hot?


    --
    Roger Hayter

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Cursitor Doom@21:1/5 to Liz Tuddenham on Mon Feb 26 10:16:41 2024
    On Mon, 26 Feb 2024 09:59:12 +0000, liz@poppyrecords.invalid.invalid
    (Liz Tuddenham) wrote:

    Cursitor Doom <cd@notformail.com> wrote:

    On Sun, 25 Feb 2024 19:16:24 +0000, liz@poppyrecords.invalid.invalid
    (Liz Tuddenham) wrote:

    Cursitor Doom <cd@notformail.com> wrote:

    On Sun, 25 Feb 2024 19:40:48 +0100, HW <none@no.no> wrote:

    On Sun, 25 Feb 2024 18:01:19 +0000, Cursitor Doom <cd@notformail.com>
    wrote:

    So - in the case of a linear power supply - the rectified output of
    the transformer typically goes straight to a big old storage cap for
    smoothing purposes. My question is: how much ripple should I expect to >> >> >>see across that cap if all's working well? I have to say I'm seeing a >> >> >>*lot* of ripple on this cap

    Check if a diode in the bridge rectifier has failed.

    Good idea! I must admit I'd overlooked that possibility. I should have
    checked the frequeny of the ripple whilst I was at it.

    The shape of the ripple will tell you if that has happened or if one of
    the diodes haas gone open circuit. Another possibility is a fault in
    some other part of the circuit that is drawing too much current - you
    might smell something getting hot.

    There's nothing getting hot, but the ripple had a definite saw-tooth
    shape to it, which could be indicative of abnormal operation.
    Shame there are no schematics available anywhere on the net; always
    makes troubleshooting much harder.

    Is the rectifier circuit half-wave or full-wave? If the latter, is the >ripple also full wave?

    Marconi have used these things (not these exact components of course,
    but the same footprint/pinout:

    http://tinyurl.com/5ab3fd7h

    So full wave. Is the ripple full wave? Well, I did mean to check the
    frequency of it, but other stuff got in the way. I'm going to have to re-investigate a bit more thoroughly later in the week.

    Some designers allowed a larger amount of ripple on the first capacitor
    to reduce the heating of the mains transformer (the copper losses are >proportional to I squared, so by allowing a longer charging time per
    cycle at a lower current, they achieved lower overall resistance
    losses). As long as the lowest part of the ripple was sufficiently far
    above the stabilised output, a series stabiliser could cope with it.

    Is the 'stabilised' line actually stable or is there ripple on it in the
    form of a series of little notches?

    Not sure what you mean by stabilised line. There are two secondary
    windings on the mains transformer which go into separate bridge
    rectifiiers of the type linked to above and then on downstream to the
    first tank caps. One line has a ton of ripple on it, the other very
    little indeed. Again, I'll need to go back and pay more attention to
    what's going on here. I've missed out too many important details.

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Cursitor Doom@21:1/5 to Roger Hayter on Mon Feb 26 11:22:21 2024
    On 26 Feb 2024 10:51:03 GMT, Roger Hayter <roger@hayter.org> wrote:

    On 26 Feb 2024 at 10:16:41 GMT, "Cursitor Doom" <cd@notformail.com> wrote:

    On Mon, 26 Feb 2024 09:59:12 +0000, liz@poppyrecords.invalid.invalid
    (Liz Tuddenham) wrote:

    Cursitor Doom <cd@notformail.com> wrote:

    On Sun, 25 Feb 2024 19:16:24 +0000, liz@poppyrecords.invalid.invalid
    (Liz Tuddenham) wrote:

    Cursitor Doom <cd@notformail.com> wrote:

    On Sun, 25 Feb 2024 19:40:48 +0100, HW <none@no.no> wrote:

    On Sun, 25 Feb 2024 18:01:19 +0000, Cursitor Doom <cd@notformail.com> >>>>>>> wrote:

    So - in the case of a linear power supply - the rectified output of >>>>>>>> the transformer typically goes straight to a big old storage cap for >>>>>>>> smoothing purposes. My question is: how much ripple should I expect to >>>>>>>> see across that cap if all's working well? I have to say I'm seeing a >>>>>>>> *lot* of ripple on this cap

    Check if a diode in the bridge rectifier has failed.

    Good idea! I must admit I'd overlooked that possibility. I should have >>>>>> checked the frequeny of the ripple whilst I was at it.

    The shape of the ripple will tell you if that has happened or if one of >>>>> the diodes haas gone open circuit. Another possibility is a fault in >>>>> some other part of the circuit that is drawing too much current - you >>>>> might smell something getting hot.

    There's nothing getting hot, but the ripple had a definite saw-tooth
    shape to it, which could be indicative of abnormal operation.
    Shame there are no schematics available anywhere on the net; always
    makes troubleshooting much harder.

    Is the rectifier circuit half-wave or full-wave? If the latter, is the
    ripple also full wave?

    Marconi have used these things (not these exact components of course,
    but the same footprint/pinout:

    http://tinyurl.com/5ab3fd7h

    So full wave. Is the ripple full wave? Well, I did mean to check the
    frequency of it, but other stuff got in the way. I'm going to have to
    re-investigate a bit more thoroughly later in the week.

    Some designers allowed a larger amount of ripple on the first capacitor
    to reduce the heating of the mains transformer (the copper losses are
    proportional to I squared, so by allowing a longer charging time per
    cycle at a lower current, they achieved lower overall resistance
    losses). As long as the lowest part of the ripple was sufficiently far
    above the stabilised output, a series stabiliser could cope with it.

    Is the 'stabilised' line actually stable or is there ripple on it in the >>> form of a series of little notches?

    Not sure what you mean by stabilised line. There are two secondary
    windings on the mains transformer which go into separate bridge
    rectifiiers of the type linked to above and then on downstream to the
    first tank caps. One line has a ton of ripple on it, the other very
    little indeed. Again, I'll need to go back and pay more attention to
    what's going on here. I've missed out too many important details.

    Older practice, when active devices were expensive and bulky was to have an >active stabiliser circuit (which also got rid of the ripple) directly after >the power supply. Nowadays they tend to have voltage regulators for each >module or circuit, local to that module. If you have the latter arrangement I >would be looking for ripple on the output of each local regulator rather than >the main power supply. Or the high ripple could be due to excessive current >being drawn from one power supply. Anything getting excessively hot?

    I don't think anything's getting unduly warm, let alone hot. But I
    can't be sure as access is impeded and they've used the most
    *enormous* finned cast aluminium heatsink for only four TO-220 package regulators. This sig gen is the 10k to 5.4Ghz one and is about 25
    years old I would guess.
    I agree with your approach on checking each regulator individually but
    the access issue is a real pain here. They clearly haven't given much
    thought to servicing with this model - and that's unusual for a
    company like Marconi/Aeroflex (or whatever they call themselves now).
    Plus the fact that they've never released schematics doesn't help
    much either.

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Liz Tuddenham@21:1/5 to Cursitor Doom on Mon Feb 26 11:15:22 2024
    Cursitor Doom <cd@notformail.com> wrote:

    On Mon, 26 Feb 2024 09:59:12 +0000, liz@poppyrecords.invalid.invalid
    (Liz Tuddenham) wrote:

    Cursitor Doom <cd@notformail.com> wrote:

    On Sun, 25 Feb 2024 19:16:24 +0000, liz@poppyrecords.invalid.invalid
    (Liz Tuddenham) wrote:

    Cursitor Doom <cd@notformail.com> wrote:

    On Sun, 25 Feb 2024 19:40:48 +0100, HW <none@no.no> wrote:

    On Sun, 25 Feb 2024 18:01:19 +0000, Cursitor Doom <cd@notformail.com> >> >> >wrote:

    So - in the case of a linear power supply - the rectified output of >> >> >>the transformer typically goes straight to a big old storage cap for >> >> >>smoothing purposes. My question is: how much ripple should I expect to
    see across that cap if all's working well? I have to say I'm seeing a >> >> >>*lot* of ripple on this cap

    Check if a diode in the bridge rectifier has failed.

    Good idea! I must admit I'd overlooked that possibility. I should have >> >> checked the frequeny of the ripple whilst I was at it.

    The shape of the ripple will tell you if that has happened or if one of >> >the diodes haas gone open circuit. Another possibility is a fault in
    some other part of the circuit that is drawing too much current - you
    might smell something getting hot.

    There's nothing getting hot, but the ripple had a definite saw-tooth
    shape to it, which could be indicative of abnormal operation.
    Shame there are no schematics available anywhere on the net; always
    makes troubleshooting much harder.

    Is the rectifier circuit half-wave or full-wave? If the latter, is the >ripple also full wave?

    Marconi have used these things (not these exact components of course,
    but the same footprint/pinout:

    http://tinyurl.com/5ab3fd7h

    So full wave. Is the ripple full wave? Well, I did mean to check the frequency of it, but other stuff got in the way. I'm going to have to re-investigate a bit more thoroughly later in the week.

    Some designers allowed a larger amount of ripple on the first capacitor
    to reduce the heating of the mains transformer (the copper losses are >proportional to I squared, so by allowing a longer charging time per
    cycle at a lower current, they achieved lower overall resistance
    losses). As long as the lowest part of the ripple was sufficiently far >above the stabilised output, a series stabiliser could cope with it.

    Is the 'stabilised' line actually stable or is there ripple on it in the >form of a series of little notches?

    Not sure what you mean by stabilised line. There are two secondary
    windings on the mains transformer which go into separate bridge
    rectifiiers of the type linked to above and then on downstream to the
    first tank caps. One line has a ton of ripple on it, the other very
    little indeed. Again, I'll need to go back and pay more attention to
    what's going on here. I've missed out too many important details.

    I was assuming there would have been some sort of stabiliser after the capacitor, so that the line downstream of it was reasonably stable and
    quiet. Alternatively they might have used a series choke (or resistor)
    and further capacitor to smoothout the ripple without actually
    stabilising the voltage

    If this is made by Marconi, the circuits for a lot of Marconi
    instruments are available online but you may need to do a bit of
    searching-out. If you can get hold of a handbook, that should always
    have the circuit in it. What is the "TF" number of this equipment?

    ["TF" is Marconispeak for "Test Fixture" i.e. any sort of test gear.]


    --
    ~ Liz Tuddenham ~
    (Remove the ".invalid"s and add ".co.uk" to reply)
    www.poppyrecords.co.uk

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From ehsjr@21:1/5 to Cursitor Doom on Tue Feb 27 17:57:32 2024
    On 2/25/2024 1:01 PM, Cursitor Doom wrote:
    So - in the case of a linear power supply - the rectified output of
    the transformer typically goes straight to a big old storage cap for smoothing purposes. My question is: how much ripple should I expect to
    see across that cap if all's working well? I have to say I'm seeing a
    *lot* of ripple on this cap in a Marconi RF signal generator and it's
    finding its way to the RF output, which of course should ideally have
    zero ripple on it. The aforementioned cap tests fine WRT ESR and
    capacitance in-circuit, I should add.

    Thanks,
    CD.


    Prior to the regulator circuit, the ripple voltage
    depends upon current drawn for a given supply at a
    given frequency. The higher the current drawn, the
    greater the ripple for your supply. So one possibility
    in your generator is that too much current is being
    drawn i.e. a defect in some circuit "downstream" of
    the bridge/cap.

    Depending on what follows the diode/capacitor supply you
    may have a regulator failure. Regulators hold the voltage
    at some specific value, thus they remove ripple. Voltage
    measurements usually reveal which regulator is failing to
    provide the proper output voltage. Again a defect downstream
    may pull the voltage down. Without a manual/schematic its
    harder to diagnose - but it is still possible. And I read
    somewhere that manuals are available. See the url posted
    below - it may help.

    Next - perhaps the ripple really isn't excessive. By that
    I mean it may be within spec for your generator. You may
    find this site helpful:
    https://groups.io/g/Marconi-Test-Instruments

    Good luck. Let us know whe you fix it!
    Ed

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Cursitor Doom@21:1/5 to ehsjr on Wed Feb 28 01:00:29 2024
    On Tue, 27 Feb 2024 17:57:32 -0500, ehsjr <ehsjr@verizon.net> wrote:

    On 2/25/2024 1:01 PM, Cursitor Doom wrote:
    So - in the case of a linear power supply - the rectified output of
    the transformer typically goes straight to a big old storage cap for
    smoothing purposes. My question is: how much ripple should I expect to
    see across that cap if all's working well? I have to say I'm seeing a
    *lot* of ripple on this cap in a Marconi RF signal generator and it's
    finding its way to the RF output, which of course should ideally have
    zero ripple on it. The aforementioned cap tests fine WRT ESR and
    capacitance in-circuit, I should add.

    Thanks,
    CD.


    Prior to the regulator circuit, the ripple voltage
    depends upon current drawn for a given supply at a
    given frequency. The higher the current drawn, the
    greater the ripple for your supply. So one possibility
    in your generator is that too much current is being
    drawn i.e. a defect in some circuit "downstream" of
    the bridge/cap.

    Depending on what follows the diode/capacitor supply you
    may have a regulator failure. Regulators hold the voltage
    at some specific value, thus they remove ripple. Voltage
    measurements usually reveal which regulator is failing to
    provide the proper output voltage. Again a defect downstream
    may pull the voltage down. Without a manual/schematic its
    harder to diagnose - but it is still possible. And I read
    somewhere that manuals are available. See the url posted
    below - it may help.

    Next - perhaps the ripple really isn't excessive. By that
    I mean it may be within spec for your generator. You may
    find this site helpful:
    https://groups.io/g/Marconi-Test-Instruments

    Good luck. Let us know whe you fix it!
    Ed

    Many thanks for that, Ed. :)