    From Bill Lamb@21:1/5 to All on Wed Oct 6 13:40:45 2021
    We have an ultrasound machine that displays the Accession Number when you query the Worklist. When you scan the patient and send the images to the PACS the Accession Number and Description do not come over. What is the reason this is happening?

    Thank You,
    Bill

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Markus Sabin@21:1/5 to billl...@gmail.com on Wed Oct 6 23:18:17 2021
    Let me make an educated guess.

    Our company offers a consulting service for testing DICOM interfaces. Mainly modality manufacturers are doing this with us. Not transferring the Accession Number from worklist results to acquired images is something that we find in almost every project.
    It seems that there is a very common misconception of what a DICOM type 2 attribute is.

    What it is: "You may leave it empty if you do not know a correct value for it" -> explicitly conveying the information "I do not have that information"
    How many implementors read it: "You may leave it empty if you want"

    Obviously, when the modality receives the Acc# via worklist, it cannot claim not to have the information, but this seems to be easy to miss.

    I would bet one box of finest franconian beer (shipping not included) that it is a bug in the modality. :-) If you would like assistance in verifying and obtaining evidence for this, you are welcome to PM me.

    Best Regards,

    Markus

    billl...@gmail.com schrieb am Mittwoch, 6. Oktober 2021 um 22:40:47 UTC+2:
    We have an ultrasound machine that displays the Accession Number when you query the Worklist. When you scan the patient and send the images to the PACS the Accession Number and Description do not come over. What is the reason this is happening?

    Thank You,
    Bill

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Bill Lamb@21:1/5 to marku...@gmail.com on Thu Oct 7 08:09:59 2021
    On Thursday, October 7, 2021 at 12:18:19 AM UTC-6, marku...@gmail.com wrote:
    Let me make an educated guess.

    Our company offers a consulting service for testing DICOM interfaces. Mainly modality manufacturers are doing this with us. Not transferring the Accession Number from worklist results to acquired images is something that we find in almost every project.
    It seems that there is a very common misconception of what a DICOM type 2 attribute is.

    What it is: "You may leave it empty if you do not know a correct value for it" -> explicitly conveying the information "I do not have that information"
    How many implementors read it: "You may leave it empty if you want"

    Obviously, when the modality receives the Acc# via worklist, it cannot claim not to have the information, but this seems to be easy to miss.

    I would bet one box of finest franconian beer (shipping not included) that it is a bug in the modality. :-) If you would like assistance in verifying and obtaining evidence for this, you are welcome to PM me.

    Best Regards,

    Markus
    billl...@gmail.com schrieb am Mittwoch, 6. Oktober 2021 um 22:40:47 UTC+2:
    We have an ultrasound machine that displays the Accession Number when you query the Worklist. When you scan the patient and send the images to the PACS the Accession Number and Description do not come over. What is the reason this is happening?

    Thank You,
    Bill
    Can you tell me how we can fix the problem. It seems as though it is not using the Study UID from the worklist, but the ultrasound is creating it's own UID, therefore the Accession Number is not coming over.

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Bill Lamb@21:1/5 to Bill Lamb on Thu Oct 7 08:45:47 2021
    On Thursday, October 7, 2021 at 9:10:01 AM UTC-6, Bill Lamb wrote:
    On Thursday, October 7, 2021 at 12:18:19 AM UTC-6, marku...@gmail.com wrote:
    Let me make an educated guess.

    Our company offers a consulting service for testing DICOM interfaces. Mainly modality manufacturers are doing this with us. Not transferring the Accession Number from worklist results to acquired images is something that we find in almost every
    project. It seems that there is a very common misconception of what a DICOM type 2 attribute is.

    What it is: "You may leave it empty if you do not know a correct value for it" -> explicitly conveying the information "I do not have that information"
    How many implementors read it: "You may leave it empty if you want"

    Obviously, when the modality receives the Acc# via worklist, it cannot claim not to have the information, but this seems to be easy to miss.

    I would bet one box of finest franconian beer (shipping not included) that it is a bug in the modality. :-) If you would like assistance in verifying and obtaining evidence for this, you are welcome to PM me.

    Best Regards,

    Markus
    billl...@gmail.com schrieb am Mittwoch, 6. Oktober 2021 um 22:40:47 UTC+2:
    We have an ultrasound machine that displays the Accession Number when you query the Worklist. When you scan the patient and send the images to the PACS the Accession Number and Description do not come over. What is the reason this is happening?

    Thank You,
    Bill
    Can you tell me how we can fix the problem. It seems as though it is not using the Study UID from the worklist, but the ultrasound is creating it's own UID, therefore the Accession Number is not coming over.
    The Images are now coming over with the Accession Number since the Manufacturer did a software upgrade to the machine. The images are coming over with a green background and pink images. Is this a Compression issue.

    --- SoupGate-Win32 v1.05
    * Origin: fsxNet Usenet Gateway (21:1/5)
  • From Bill Lamb@21:1/5 to Bill Lamb on Thu Oct 7 10:08:08 2021
    On Thursday, October 7, 2021 at 9:45:49 AM UTC-6, Bill Lamb wrote:
    On Thursday, October 7, 2021 at 9:10:01 AM UTC-6, Bill Lamb wrote:
    On Thursday, October 7, 2021 at 12:18:19 AM UTC-6, marku...@gmail.com wrote:
    Let me make an educated guess.

    Our company offers a consulting service for testing DICOM interfaces. Mainly modality manufacturers are doing this with us. Not transferring the Accession Number from worklist results to acquired images is something that we find in almost every
    project. It seems that there is a very common misconception of what a DICOM type 2 attribute is.

    What it is: "You may leave it empty if you do not know a correct value for it" -> explicitly conveying the information "I do not have that information"
    How many implementors read it: "You may leave it empty if you want"

    Obviously, when the modality receives the Acc# via worklist, it cannot claim not to have the information, but this seems to be easy to miss.

    I would bet one box of finest franconian beer (shipping not included) that it is a bug in the modality. :-) If you would like assistance in verifying and obtaining evidence for this, you are welcome to PM me.

    Best Regards,

    Markus
    billl...@gmail.com schrieb am Mittwoch, 6. Oktober 2021 um 22:40:47 UTC+2:
    We have an ultrasound machine that displays the Accession Number when you query the Worklist. When you scan the patient and send the images to the PACS the Accession Number and Description do not come over. What is the reason this is happening?

    Thank You,
    Bill
    Can you tell me how we can fix the problem. It seems as though it is not using the Study UID from the worklist, but the ultrasound is creating it's own UID, therefore the Accession Number is not coming over.
    The Images are now coming over with the Accession Number since the Manufacturer did a software upgrade to the machine. The images are coming over with a green background and pink images. Is this a Compression issue.
    I changed the compression to uncompressed and the images come over just fine.